More on A.pellucida

Thread view: 
Enable EMail Alerts  Start New Thread
Thread rating: 
NoSpam - 20 Jan 2007 19:05 GMT
Microscope Group,

I wish to present a light microscope image (LM) of A. pellucida,
with a determination of the distance between the striae. The re-
sult is compared with those obtained on various A. pellucida
frustules imaged by field emission microscopy (FESEM). Dr. G.
Gaugler, USA did this work and his SEM images are presented at
his site, www.photoweb.net/micro/ .

My LM image is at
www.mikroskopie-forum.de/read.php?2,25394,25394#msg-25394 .

Pertinent details are as follows. The sample for the FESEM
images was specially prepared by Mr. Klaus Kemp, UK. It con-
sisted of nine A.pellucida frustules placed on top of a coverglass
in free air, no mounting medium was used. Dr. Gaugler used
Palladium to sputter the sample before doing SEM microscopy.

The frustule used for the LM picture is on one of Mr. Kemp's
test slides. It is mounted the conventional way in Zrax and
there is one A. pellucida diatom together with samples of other
species on the test slide.

All of the samples, SEM and LM show "inner views" of the
diatom, that is views with the concave side poining up.
Mr. Klaus Kemp's site is at www.diatoms.co.uk/ .

The LM picture was obtained using an Olympus BX51 micros-
cope using a planAPO objective 100x/1.35 using their universal
condenser with top oil lens and double immersion. DIC was
used to enhance contrast and the image was projected with
a 2.5 Olympus eyepiece upon the CCD of a Pentax *ist DS
digital, single lens reflex camera. The vertical dimension of the
CCD is 1.57 cm. JPG format was used with 2000 pixels along
the vertical. Illumination was by electronic flash and image
processing was limited to adjustment by Auto Levels, follow-
ed by Grayscale in Adobe Photoshop. No other processing
was done. The image horizontal extent was shortened, the
vertical dimension preserved at 1980 pixels from the full 2000
and that is what is shown at
www.mikroskopie-forum.de/read.php?2,25394,25394#msg-25394 .

While I have been able to resolve the striae I was unable to re-
solve the striae into punctae (dots) using my light microscope.

Since I do not have a stage micrometer I determined the distance
between the striae as follows: I used the magnification of
the objective of 100x and that of the photoeyepiece of 2.5x to
calculate a total magnification of 250 x. Counting the number
of striae in 150 pixels as 18, one obtains 150/18 = 8.33 pixels
per stria. Each pixel represents 1.57cm/2000 = 7.85 micrometers
on the CCD and with a magnification of 250 this corresponds
for 8 pixels to 8.33 x 7.85/250 = 0.26 micrometers for the distance
between striae.

Dr. Gary Gaugler used his Zeiss Supra 55VP FESEM instrument
to image and measure important distance parameters on the
samples for SEM microscopy. One needs to look at the images
at his site www.photoweb.net/micro/  to appreciate the signifi-
cance of the results. An overview showing a bird's eye view of
the samples is given under the thumb nail Apellucida-specimens.

Currently there are details for five of the nine A. pellucida dia-
toms posted. This allows a comparison of variability between
the parameters for individuals belonging to the same species.
It will be seen that the variability of distances between striae and
between the pores of a stria (dots) amounts to as much as 9 %.
This is probably more than has generally been expected. Please
refer to Apellucida-50kx-2.jpg, Apellucida-50kx-D.jpg,
Apellucida-50kx-E.jpg, Apellucida-50kx-F.jpg and
Apellucida-50kx-I.jpg for images and measurements.

For these diatoms the following results apply (all in micrometers):
-2 distance between striae 0.289  distance between dots 0.229
-D                                               0.272
0.176
-E                                                0.301
0.283
-F                                                0.284
0.181
-G                                                0.275
0.161.
For LM from above:
LM                                              0.26
...........................

Even within the same diatom distances between pores may vary.
This is best appreciated at high magnifications. Also note the
thickness measurements of the frustule's bottom (Apellucida-
100kx-3). These were made at 25 degrees inclination. If a frustule
is inclined, the bottom layer may obstruct a fair portion of the pore
and have an effect upon the LM picture. The measured thickness
of the bottom was about 0.09 micrometers at the center and less
at the ends.

It will also be noted that some pores are plugged up, which might
make LM observation of details more difficult on some diatoms.
Particularly diatom E shows many plugged pores.

Now the question may be raised whether separation of the pores
within a stria, can be done by LM. I do not believe that I can do it
with the setup described above. I might be possible to do better
with a projection magnification of  5x or more.

However Peter Hoebel at
www.mikroskopie-forum.de/read.php?2,25384,25384#msg-25384
presents such a LM picture with dots separated for A. pellucida. He
had used a near UV LED and averaged 1000 images to obtain the
separate dots. Details can be found at the above link.

Regards
G.R.
rene - 22 Jan 2007 12:13 GMT
You shouldn't use DIC for ultimate resolution.

Ren?.

> Now the question may be raised whether separation of the pores
> within a stria, can be done by LM. I do not believe that I can do it
> with the setup described above. I might be possible to do better
> with a projection magnification of  5x or more.
NoSpam - 22 Jan 2007 16:45 GMT
Rene,

You are of course correct, in principle.

The problem with A. pellucida however is not only the one
posed by resolution, but also and very importantly, by contrast.

With insufficient contrast one is unable to see the object and
with insufficient resolution one will not see details. One must
find a compromise. In my case the DIC image was far superior
to bright field or dark field. The imaging of an object like
A. pellucida ist not only a question of resolution, but of the
objects innate contrast and of the modulation transfer function of
the entire setup. The latter involves more than the NA of the
objective.

Why dont you point us to your images and describe the details
under which they were taken!

G.R.

You shouldn't use DIC for ultimate resolution.

René.

> Now the question may be raised whether separation of the pores
> within a stria, can be done by LM. I do not believe that I can do it
> with the setup described above. I might be possible to do better
> with a projection magnification of  5x or more.
rene - 22 Jan 2007 17:32 GMT
Discussion of KKemp material and imaging of Ap has been going on for
years on the yahoo microscope discussion group, you can find my images
overthere. It is similar to your result, but with an old microscope
with 15W Tungsten/420nm bluefilter, annular illumination with a
modified BF condenser and a 70 year old achromatic Beck 1.3 NA
objective.
Many others have done the same exercise with better results (then mine)
by playing with oblique illumination.
Indeed, lack of contrast is the bottleneck, caused by 'thin frustules'
(as KKemp calls it), ie little silicification between pores in the
striae. The (inter)striae themselves are heavily silicified bars,
therefore reasonably easy to image compared to the pores.

René.
NoSpam - 22 Jan 2007 18:08 GMT
Rene,

Well, thank you for your reply and the reference to the
Yahoo group.

I am aware of the long history surrounding the imaging
of A. pellucida and the many attempts, more or less success-
ful to resolve striae or dots.

What has been missing so far was a comparison of a good
light microscopy image with superb and quantitatively
precise SEM images. These were obtained by Dr. G. Gaugler
and presented in the current thread. These images are unique
in several respects: They show measurement of important
parameters on the frustules of A. pellucida, not only on one,
but on several specimens. This allows the comparison of
these values on samples of the SAME origin and of the
same size. It also shows a light microscope image of A. pel-
lucida from the SAME source and of comparable size.

I think this makes for very interesting reading for those inter-
ested in the subject and is not meant to be imply that others
have not obtained good pictures of A. pellucida. In fact I
do point to a very detailed image of A. pellucida by a link.

GR.

Discussion of KKemp material and imaging of Ap has been going on for
years on the yahoo microscope discussion group, you can find my images
overthere. It is similar to your result, but with an old microscope
with 15W Tungsten/420nm bluefilter, annular illumination with a
modified BF condenser and a 70 year old achromatic Beck 1.3 NA
objective.
Many others have done the same exercise with better results (then mine)
by playing with oblique illumination.
Indeed, lack of contrast is the bottleneck, caused by 'thin frustules'
(as KKemp calls it), ie little silicification between pores in the
striae. The (inter)striae themselves are heavily silicified bars,
therefore reasonably easy to image compared to the pores.

René.
gto - 23 Jan 2007 05:40 GMT
As Gary clearly showed with his images, one should use SEM for AP. Regarding
resolving AP with LM, what has been achieved at the beginning of last
century is all we will be able to resolve using LM until someone comes up
with a greatly improved LM setup. (Well, I am not considering recent
progress in looking at self-luminous particles at the moment.) The good old
times proofs once again to be a great source for knowledge ;-)

But, as usual, you are correct. DIC is not the best choice for AP. COL is
the champ. DF is recommended. With lambda = 420nm, one should be able to
resolve AP with almost any reasonable scope and a decent DF condenser. (Of
course, both oiled to the slide.)

Another question, do you use a water immersion lens? If yes, do you have the
one that works with a coverglass or without? If I remember correctly, you
use a LOMO at home and a Zeiss at work. - I am looking for one for my
Ortholux.

Cheers,

Gregor

You shouldn't use DIC for ultimate resolution.

René.

> Now the question may be raised whether separation of the pores
> within a stria, can be done by LM. I do not believe that I can do it
> with the setup described above. I might be possible to do better
> with a projection magnification of  5x or more.
rene - 23 Jan 2007 08:51 GMT
Hi Gregor,  I have used a CZJ for this stuff. Yes, I also have the Lomo
water immersions, particularly the 70/1.23 apo is ok for diatom work
with COL. I managed to get it working for DF with the CZJ aplanatic 1.4
condenser, by inserting the DF stop just below the frontlens (not in
the filterholder!). Haven't tried it with Ap though. Yes, it is the one
with correction collar 0.11-0.22. It's an oldy, contrast in BF is not
great, I prefer to use the achro 85/1.0 WI for normal work. I know Lomo
made a new batch of the 70x in 2004, not sure whether they managed to
improve the contrast, but for a couple of 100$ (new!) they are great
value for money.

René.

gto schreef:
> As Gary clearly showed with his images, one should use SEM for AP. Regarding
> resolving AP with LM, what has been achieved at the beginning of last
[quoted text clipped - 25 lines]
> > with the setup described above. I might be possible to do better
> > with a projection magnification of  5x or more.
Kevin Cunningham - 23 Jan 2007 14:19 GMT
> As Gary clearly showed with his images, one should use SEM for AP.
> Regarding resolving AP with LM, what has been achieved at the beginning of
[quoted text clipped - 16 lines]
>
> Gregor

DF reduces resolution over bright field or DIC.  With a DF system the NA of
the objective and condenser is quite low in comparision to bright field and
DIC.  NA is resolution, resolution is NA.  Thats why DF is yesterday's news.

Thanks,

Kevin Cunningham
SMS

> You shouldn't use DIC for ultimate resolution.
>
[quoted text clipped - 4 lines]
>> with the setup described above. I might be possible to do better
>> with a projection magnification of  5x or more.
rene - 23 Jan 2007 14:54 GMT
Is this your intuition or where did you get this from??
R.

> DF reduces resolution over bright field or DIC.  With a DF system the NA of
> the objective and condenser is quite low in comparision to bright field and
[quoted text clipped - 4 lines]
> Kevin Cunningham
> SMS
Kevin Cunningham - 24 Jan 2007 23:19 GMT
> Is this your intuition or where did you get this from??
> R.

Dr. Abbe, 1872.  Zernicke's paper and talk he delivered in 1938 at the Carl
Zeiss Wercke, Jena.  Publication by Dr. Nomarski.  Publication by Dr.
Francon.  Training from Zeiss, Nikon, Olympus and Leica.  Working in the
field since the late '70's.  In that period I sold 2 DF condensers and I
accidentally sold 6.  Thats in 35 years.  I've never sold a high power DF
objective to a researcher.  I've had the honor of working at the Rockefeller
Inst., Harvard, Yale, Stanford, US-C, UG-A, Down State Med., Emory, etc.

Rene, you do a lot of good for this group, your questions are terrific and
your responses are on point, usually.  This is not one of the usual times.

Kevin Cunningham
SMS

>> DF reduces resolution over bright field or DIC.  With a DF system the NA
>> of
[quoted text clipped - 7 lines]
>> Kevin Cunningham
>> SMS
rene - 25 Jan 2007 08:30 GMT
Kevin, yes, I might have been too sharp in this discussion, might
refllect my state of mind these months.
Anyway, like for like, a 40x BF or in DF shouldn't suffer in
resolution. I achieved DF with a 70x objective with the full aperture
of 1.23 by relatively simple means (a patched aplanatic 1.4 condenser).
The view in DF is *very* detailed, see my yahoo photo album for an
example. DF is just not very practical for most specimen, especially
not in high NA. And, I agree with Gregor, even for diatoms I prefer COL
over DF.

René.
GTO - 25 Jan 2007 01:32 GMT
> DF reduces resolution over bright field or DIC.  With a DF system the NA
> of the objective and condenser is quite low in comparision to bright field
> and DIC.  NA is resolution, resolution is NA.  Thats why DF is yesterday's
> news.

Kevin:

Yes, DF reduces resolution over BF and DIC due to the closed iris located
inside the oil immersion objective. We both know that it must be closed to
around 1.10 but not more than 1.20 in order to obtain decent darkfield
illumination. The condenser for DF does not lower resolution since any
decent DF condenser using oil immersion provides a light cone that
illuminates between NA 1.20 to 1.40. Hence the resolution decrease is purely
due to the objective's iris (or stop).

However, the problem to distinguish certain, tiny diatom structures, is not
just a lack of resolution; it is the availability of sufficient contrast. In
DF, I can detect particles that are way below the resolution limit of BF and
DIC. But these particles appear in DF larger than their physical dimensions.
It is clear from diffraction theory of light that when the size of a
particle or feature falls below the minimal particle size that can be
resolved with a given optical setup, the size of the image of said particle
is no longer smaller than the minimal particle size, but its contribution to
the image contrast decreases as the diameter of the  subresolution particle
gets smaller.

DF is recommended for this application in circles that study diatoms
intensively using LM, while DIC isn't due to its image artifacts that make
particle/feature interpretations difficult. Amateur microscopists tend to
use DIC and PC to produce pretty pictures. Personally, I prefer COL for
diatoms, which happens to decrease contrast when compared to DF but, of
course, offers the full resolution of the objective/condenser combination.

Cheers,

Gregor
Kevin Cunningham - 23 Jan 2007 14:15 GMT
You shouldn't use DIC for ultimate resolution.

René.

> Now the question may be raised whether separation of the pores
> within a stria, can be done by LM. I do not believe that I can do it
> with the setup described above. I might be possible to do better
> with a projection magnification of  5x or more.

Let me disagree with Rene.  Actually if you look at the light path diagrams
and the formal and informal papers on DIC it actually increases resolution
by a tiny amount.  This is caused by the two light path phenominon that
creates DIC.

Thanks,

Kevin Cunningham
SMS
rene - 23 Jan 2007 14:51 GMT
Not my experience. Have also never seen A.p. resolved to dots with DIC,
and especially diatoms are suposed to be ideal. Check the MTF's
http://www.olympusmicro.com/primer/java/mtf/contrasttechniques/index.html
What (in)formal papers are you talking about?

René.

Kevin Cunningham schreef:
> You shouldn't use DIC for ultimate resolution.
>
[quoted text clipped - 14 lines]
> Kevin Cunningham
> SMS
Kevin Cunningham - 24 Jan 2007 23:19 GMT
Not my experience. Have also never seen A.p. resolved to dots with DIC,
and especially diatoms are suposed to be ideal. Check the MTF's
http://www.olympusmicro.com/primer/java/mtf/contrasttechniques/index.html
What (in)formal papers are you talking about?

René.

______________________________________________________--

Rene,  An informal publication can be a talk or a letter.

Kevin Cunningham
SMS

Kevin Cunningham schreef:
> You shouldn't use DIC for ultimate resolution.
>
[quoted text clipped - 15 lines]
> Kevin Cunningham
> SMS
cladocera56@ntlworld.com - 22 Jan 2007 18:57 GMT
Hi

Thanks for sharing your A. pellucida results and as you remark the
comparison of SEM and LM photos of the same material is valuable.

As an aside, I think the work of Spitta to me as a hobbyist is jaw
dropping if not humbling. His book 'Microscopy' 1909 2nd edition has
very well resolved photos of A. pellucida to 'dots' with blue light and
convincingly resolved to 'dots' with green light, all of high contrast.
These photos are presented in conjunction with his chapter devoted to
using this and other diatom species and Abbe test plate as test
objects.

His earlier book 'Photomicrography' 1899 has two text pages on
photographing A. pellucida and a beautiful photo of this diatom
resolved to striae. Although it's his photo of another test diatom P.
angulatum in 'Microscopy' that blows me away, not a tricky one to
resolve but it's a beautifully crisp photo of this diatom with internal
detail.

with regards

David
NoSpam - 22 Jan 2007 19:44 GMT
Hallo,

Thank so much for for your constructive contribution and the
reference to Edmund J. Spittas work. I have looked at Google
Books and found "Photomicrography" of 1899 I downloaded
it and will look at the photos. I could not locate the other work
"Microscopy" of 1909. The existence of both was unknown to
me.

Regarding the ability of some observers to resolve A. pellucida
into dots while others were not able to do so, may have two causes.

One is that A. pellucida might occur with fine or coarse structure.
There is strong variability of the parameters for samples from the
same site. There may be even stronger variability for samples from
different locations. This aspect is something which Dr. Gaugler
will attempt to clarify in the near future by using some special
samples from Mr. K. Kemp.

The other cause may have to do with what is nowadays called the
modulation transfer function. Since modern images are taken
on 36 mm film or on CCD or CMOS sensors, the size of the
a resolved element thrown upon the film or sensor compared to
the film grain and pixel size is a most important parameter. It is
desirable to have as large as possible a magnification upon
the sensitive surface as is practical and combine this with as
small as practical a grain size on film or of a pixel. The latter
of course needs to have a certain size to achieve a good S/N
ratio.

Nowadays with 36 mm film and small CCD or CMOS sizes
there are limits. In the old days when microscopists used
very large photographic plates and large projection distances
these limits were more favorable.

These two causes may have a bearing while microscopists
of over one hundred years ago, were able to resolve A. pelluci-
da into dots.

Greetings
GR.

> Hi
>
[quoted text clipped - 19 lines]
>
> David
rene - 23 Jan 2007 09:59 GMT
No Spam, from my memory the first image I have is in Neuhauss, Lehrbuch
der Mikrophotographie 1907. I would have to find it again, but it was
made with 'Cadmium light', 275nm I believe was mentioned. Search for
Neuhauss in the yahoo microscope archives and you will find additional
info on earlier and later work on A.p.

 It is
> desirable to have as large as possible a magnification upon
> the sensitive surface as is practical and combine this with as
[quoted text clipped - 6 lines]
> very large photographic plates and large projection distances
> these limits were more favorable.

The more empty magnification, the lower the contrast. 36mm fim and 3Mp
cameras are more then adequate.

The problem with A.p. for test specimen is it's variability in
silification (or so I believe) and possibly variation in RI of mounting
media. Even slight differences in refraction index between batches will
have tremendous impact on contrast. This all makes a diatom test slide
a personal subjective test.

Last comment, the use of contrast enhancement techniques like
oblique/single sideband/DIC/phase contrast makes it even more
subjective. Only brightfield will show the quality of an optical
system. As you say, the resolution is there, no problem, but without
contrast (by the optical system OR the test specimen), you cannot see
it.

Oh yes, and be careful with taking sizes from SEM, in respect of tilted
valves, also SEM calibration. I wouldn't trust my own SEM pics from
A.p. within 10%. I'm sure Gary's FESEM would be better in that respect.

Ren?.
Keith Shaw - 23 Jan 2007 18:03 GMT
Hello again Rene,

You wrote:
> ..from my memory the first image I have is in Neuhauss, Lehrbuch
> der Mikrophotographie 1907. I would have to find it again, but it was
> made with 'Cadmium light', 275nm I believe was mentioned.
> Search for Neuhauss in the yahoo microscope archives

While working on the prototype archive website (some time back!), I often
wondered if the
image shown here: http://204.3.164.54/images/Pic_NH_Ap.htm
is actually Neuhauss' A.pellucida.

Best regards,  Keith
NoSpam - 23 Jan 2007 18:44 GMT
Rene,

You stated:
>The more empty magnification, the lower the contrast. 36mm fim and
3Mp cameras are more then adequate.>

This statement is incorrect. Contrast is affected by the ratio of image
detail relative to grain size or pixel size. That is: for a given size Y of
the
resolved element on a given sensitive surface, the contrast is enhanced
by increasing Y by a factor n  to n x Y, where n is larger than 1. At the
same
time the area under the image of the resolved element increases by
(n x Y)**2, while the noise increases by n x Y. This results in an improve-
ment of  S/N by a factor of n x Y. This result would even be correct
up to a certain factor of n at which resolution would be satisfactory
for the average eye, that is into the range of empty magnification.

For those who believe more in experiments than in theory I have done
the experiment: I produced an image of A. pellucida by projecting the
intermediate image of this diatom upon the CCD of my *ist Pentax (n=1)
and was UNable to resolve the striae. I then projected the intermediate
image through a 2.5 projection lens (n=2.5)and could resolve the striae
perfectly well as shown at
http://www.mikroskopie-forum.de/read.php?2,25394,25394#msg-25394 .

To forestall any questions: the intermediate image in my Olympus scope
is fully corrected and does not require additon of a projection lens
to take full advantage of the resolution the objective is capable of.

It would be nice, would it not be, if you presented the dotted image
from Neuhauss' book? I have tried and I can't get a hold of this work
and it would certainly be of interest to many to be able to see it.

GR.
Gary G - 23 Jan 2007 22:04 GMT
>Last comment, the use of contrast enhancement techniques like
>oblique/single sideband/DIC/phase contrast makes it even more
[quoted text clipped - 8 lines]
>
>René.

A coated specimen will increase LM contrast.  Done right, you won't
see the coating.  Au is not good.  Au/Pd, Ir, Pt or Pd are IMO better.
The surfaces are coated thicker than the pores due to plasma flux
variations (more difficult to get into the holes the same as the flat
surfaces).

The tilted specimens were optimal for thickness measurement.  These
specimens were the internal (concave) view rather than the external
(convex) view.  AP is like a W-shaped specimen with the middle of each
side flat on the cover slip.  So, if one considers one side, it is
like a U where the bottom of the U is flat on the cover slip.  This
allows for contrast variation between the slip and specimen.  Tricky
but doable.

I use Geller MRS-5 microscopy calibration standard but only up to
500KX.  The standard has patterns at 80nm (+- 1nm) on up to 3um.  Most
of my work is between 5KX and 300KX.  Since calibration correction is
electronic, it is easy to check and tweak if necessary.  Typical
calibration limits are around +- 1% of FS at high mag and about 2-4%
at low mag (due to inability to put the cursors exactly on the edges).
So, this is mostly an operator problem.

gg

Kiss French.  Drink California.
rene - 25 Jan 2007 12:30 GMT
Hi Gary,

I've tried goldsputtering only, in different thickness on sectors of
the same coverglass. A thin layer increased visibility of the striae in
BF, no sign of the pores. In annular illumination (COL), pores were
visible in the UNcoated part of the slide, not in the coated parts. Not
sure why, but I can think of a couple of reasons to do with annular
illumination.
It was an experiment after I saw metallized slides of Ap, I think
aluminium shadowcasts. That created an image that resembled
phasecontrast. The striae were visible in BF, much better then my
goldsputtered specimen, no doubt because of the shadowing technique
used. No sign of the pores, alas.

René..
Gary G - 25 Jan 2007 18:09 GMT
>Hi Gary,
>
[quoted text clipped - 11 lines]
>
>René..

Au is good for low mag images only.  Even when ideally laid down, and
not thick, it forms a spider web of Au rather than an amorphous layer.
The next option is Au/Pd which if done right will work up to about
200KX.  At this point, the coating becomes quite visible as mounds.
High vacuum low current coating with Pt, Ir or Pd produces extremely
fine grain results that are not seen up to at least 600KX.  Probably
the best coating is Cr plasma deposition.  However, Cr oxidizes
rapidly and thus, tends to be useful only for a session.

gg

Kiss French.  Drink California.
Keith Shaw - 24 Jan 2007 16:51 GMT
I was incorrect re identification of the A.pellucida image referenced in a
previous post, i.e. it may have been by Neuhauss.

On going back over some old notes, it appears that this image is from
W.B.Carpenter's book  "The Microscope and its Revelations" published in
1891. It is accredited to Prof. Van Heurck, a rather well known
microscopist!

This image most probably originated with Martin Mauch at :
www.baertierchen.de

Regardless, the definition of the dots within the right-hand section of
this mount is startling - in fact it's hard to believe.

It would be real interesting if anyone can dig out information on the
equipment and illumination technique used by Van Heurck.
My notes indicate a Zeiss objective (presumably 100x) with NA 1.4 - a very
nice piece of glass (hand ground) back in the 1890s.

-Keith
rene - 24 Jan 2007 17:27 GMT
Not entirely sure, might have been the 1.6 NA (!!) exotic immersion
one, something like monobroomnaphtalene immersion directly on an
uncovered specimen. Will dig out the Neuhauss book tonight.

René.

> I was incorrect re identification of the A.pellucida image referenced in a
> previous post, i.e. it may have been by Neuhauss.
[quoted text clipped - 15 lines]
>
> -Keith
rene - 25 Jan 2007 11:50 GMT
I have found some more information in an article by Trivelli and Lincke
(1931):
Fraenkel and Pfeiffer showed photomicrographs of the Ap dots first in
1889 (according to Neuhauss). In 1890 van Heurck and 1893 Zettnow
succeeded also with use of the new Zeiss apo 1.6NA, as said before a
monobrome naphtalene immersion lens.
Zettnow was able to count dots to 52/10um. He used a filter of
cuppramonium solution and iodine (ie violet). The diatom was mounted in
mercury iodide, and as said before the 1.6NA lens was a monobrome
naphtalene immersion lens.

Neuhauss in 1907 was slightly contradictonary as he mentioned van
Heurck and Zettnow used the same lens, but a Realgar mounted slide and
blue light (violet would be absorbed by Realgar). I guess both will
work.

Watch this remark on van Heurck's images (Neuhauss p 264, translated):
van Heurck tried to resolve the striae (Langstreifung) and pores
(Querstreifung) (1886). However, he masked in many images along the
edge of the diatom. Therefore, it was impossible to differentiate the
'Querstreifung' from diffraction lines [that are optical artefacts from
the edge of the valve].

Neuhauss reproduced a truly magnificent image from Ap, made with 275nm
UV-light in 1907. As far as I can make out this image originate from
the Zeiss labs where the new UV monochromat was developed, a glycerin
immersion 1.25NA. If Keith contacts me I can send him a couple of
images from the page of this book so that everybody could see it on his
website, as I have no webspace to show it on.

René.
Keith Shaw - 25 Jan 2007 22:00 GMT
> Neuhauss reproduced a truly magnificent image from Ap, made with 275nm
> UV-light in 1907.
> ..
> If Keith contacts me I can send him a couple of  images from the page of
> this book
> so that everybody could see it on his website...

Courtesy of Rene, here they are:
 http://www.kscitech.com/Pic_04.1.htm
 http://www.kscitech.com/Pic_04.2.htm

Enjoy!!
Gary G - 26 Jan 2007 04:39 GMT
>> Neuhauss reproduced a truly magnificent image from Ap, made with 275nm
>> UV-light in 1907.
[quoted text clipped - 8 lines]
>
>Enjoy!!

If I had images like these with my SEM I would be very upset.

The wavelength of an electron beam is less than .5A and is obviously
superior to LM.  Duh.  I know this.  The same problems occur with SEM
as with LM....astigmatism and distortion.  So, LM and SEM have limits.
With LM, the variables are known. With SEM, the variables are mostly
confined to the column.  Thus, much of perfection is lost in trying to
get a perfect beam.  Not possible.  But Zeiss and others make huge
investments in this area.

gg

Kiss French.  Drink California.
NoSpam - 24 Jan 2007 20:09 GMT
Dear Keith,

Yes! The image is by Dr. Henri van Heurck. The image and des-
criptive details can be seen at
http://www.baertierchen.de/archiv.html ,
which is the site of the Baertierchen microscopist Martin Mach
of Munich, Germany.

Mr. Martin Mach gives Willliam B. Carpenter "The Microscope"
1891, as the source of this image of A. pellucida.

Actually William B. Carpenter's "The Microscope and its
Revelations" is a series of books with this title which contain
important work in the fields of microscopy. The mentioned volume
is edited by W.H. Dallinger, since W.Carpenter had died previous
to the publication of this volume.

The book mentioned by Mr. Mach can be found at the Google book
site under
http://books.google.com/books?vid=OCLC00999815&id=YIYAAAAAMAAJ&pg=RA1-PR10&l
pg=RA1-PR10&dq=William+carpenter++microscopy+1891#PRA1-PR21,M1
and may be downloaded as a pdf file without charge.

I shall briefly describe what I found giving the page numbers
which the reader can enter on the bottom of the pdf reader.
This will save time because the search function does not work
on this type of pdf file.

Choose p. 13 on the counter on the lowest line of the Adobe pdf reader
and you will see the Frontispiece. Fig. 5 of the frontispiece shows:
"Amphipleura pellucida  x 1860 diams., by apochromatic 1/8 1.4 N.A.
illuminated by a very oblique pencil in one azimuth along the valve."

Choose p. 25 and you will get to read about Plate XI which contains
three images of A. pellucida at magnifications of 2000x, 3000x and
2600x. The description goes into very interesting details of what Dr.
van Heurck thought was the structure of the valves of this diatom.
The text is too extensive to reproduce here.

Unfortunately I could not find Plate XI in the Google pdf file. Only
Plate I seems to have been scanned.

There are additional descriptions and remarks regarding the microscopy
of A. pellucida. You will find them on pp. 90, 106, 116, 260 and 551,
again using the search function at the bottom of the Adobe reader.

Page 551 makes very interesting statements regarding the great varia-
bility of the structure of A. pellucida which can present with easily
resolved valve details or details which are extremely difficult to
resolve. Again I must refer to the original text. These statements,
if supported by modern data, may have an important bearing on the
ability to resolve this diatom into "dots" using light microscopy.

Dr. Gary Gaugler will most likely explore the variability of
A. pellucida using his FESEM. Mr. Klaus Kemp (UK) will supply
samples.

Finally I wish to point out that my image of A. pellucida (see
the link in my original post) was obtained using a magnification
of 250x (100x from the objective, 2.5x from the projection lens),
while Dr. von Heurck used magnifications of 1800x to 3000x.

GR.

> I was incorrect re identification of the A.pellucida image referenced in a
> previous post, i.e. it may have been by Neuhauss.
[quoted text clipped - 16 lines]
>
> -Keith
rene - 25 Jan 2007 12:22 GMT
Hi NoSpam, can you tell me where exactly this is located on the
Baertierchen site??
Also, you seem to have been lucky to be able to download Spitta's and
Carpenter/Dallinger's book, I could only get snippets which were
useless. Would you mind sending them to www.yousendit.com, so that we
could download them from there?

In return some more comments on your messages:

You stated:

>The more empty magnification, the lower the contrast. 36mm fim and
>3Mp cameras are more then adequate.

I stand with this conclusion. A 100x 1.4NA lens with 20mm secundary
image (such as seen with a 10x eyepiece with fieldnumber 20) only needs
2000 pixels along the middle line for all detail for all to be resolved
(ie 2 pixels for 0.2 um). If you fix a 36mm frame inside the field of
view, you will need less pixels, but add some to compensate for the
Bayer filter in the digicam, then you will have enough with 3Mp.
I agree with you, increase pixelSIZE (but not total amount of pixels)
and with corresponding extra magnification will reduce noise, but that
is irrelevant in this discussion.

> I produced an image of A. pellucida by projecting the
>intermediate image of this diatom upon the CCD of my *ist Pentax (n=1)
>and was UNable to resolve the striae. I then projected the intermediate
>image through a 2.5 projection lens (n=2.5)and could resolve the striae
>perfectly well as shown at
>http://www.mikroskopie-forum.de/read.php?2,25394,25394#msg-25394 .

I'm not sure exactly where your exp did go wrong. By 'direct
projection'  you mean you simply took out the projection lens/eyepiece
and adjusted focus? That would increase the effective tubelength a
couple of cm from the original 160mm it was designed for. This would
create spherical abberation, ie unsharpness. Your CCD should be at the
place where the secundary image is placed (where your eyepiece 'picks
it up'), somewhere a cm or so below the rim of the tube).

> Finally I wish to point out that my image of A. pellucida (see
> the link in my original post) was obtained using a magnification
> of 250x (100x from the objective, 2.5x from the projection lens),
> while Dr. von Heurck used magnifications of 1800x to 3000x.

That would make your specimen a very big one ;-)

René.

ps, keep it up, you are learning a lot.
NoSpam - 25 Jan 2007 17:28 GMT
Rene,

I am beginning to have doubts regarding your competence.
Your recommendation that I should keep it up because I will be
learning a lot is entirely misplaced as the following lines will
clearly show.

From your recent post:
>"rene" <renevanwezel@hotmail.com> wrote in message >news:1169727731.244566.22640@k78g2000cwa.googlegroups.com...
>Hi NoSpam, can you tell me where exactly this is located on the
>Baertierchen site??

The site where you can find the article by Martin Mach showing the
van Heurck A. pellucida picture is under
http://www.baertierchen.de/archiv.html
where there is the telling title: Februar 2004: Feine Strukturen.

From your recent post:
>Also, you seem to have been lucky to be able to download Spitta's and
>Carpenter/Dallinger's book, I could only get snippets which were
>useless. Would you mind sending them to www.yousendit.com, so that
>we could download them from there?

The Carpenter/Dallinger book can be found exactly where I
said it can be found. Now you will know that long, that is
several line long links, are not properly handled in newsnet
messages. It is therefore unproductive to just click on the
address as given. What you need to do is copy the entire link
to a text editor, like Notepad. You do this by "copy" the link
in sections, line by line, and reconstructing the address
by "paste" in your text editor. Then you copy the entire address
so obtained into your browser and VOILA, there is the book.
(To be entirely sure I repeated this process just now and it
works, like it always does.) I do not know how else I can make
it easy to find this source. Believe me, I spent a lot more time
finding it and then the page numbers, than it would have taken
you to follow the described process! So go ahead, download
the book, look at the pages I quoted and you will be able to
read the text and see the picture.

You are quoting from a prior text of yours:
>>The more empty magnification, the lower the contrast. 36mm fim and
>>3Mp cameras are more then adequate.
Your addition to that text is:
>I stand with this conclusion. A 100x 1.4NA lens with 20mm secundary
>image (such as seen with a 10x eyepiece with fieldnumber 20) only needs
[quoted text clipped - 5 lines]
>and with corresponding extra magnification will reduce noise, but that
>is irrelevant in this discussion.

You seem confused by the issue of contrast. I will say it
one more, final time: contrast increases when the target's
image spatial frquencies move to the left on the MTF curve.
The only way for this to happen at a fixed pixel or grain size is
to decrease the spatial frequency, that is to increase the size of
the image.

What may lead you astray is the custom of presenting the MTF
curve for targets with 100 % contrast. The x% which are deemed
sufficient for detection are then at a certain position on the hori-
zontal axis. BUT, when the contrast of the target is less than 100%,
this point moves to the left and for small contrast it moves a lot to
the left.

Your quote from a prior post of mine:
>>I produced an image of A. pellucida by projecting the
>>intermediate image of this diatom upon the CCD of my *ist Pentax (n=1)
>>and was UNable to resolve the striae. I then projected the intermediate
>>image through a 2.5 projection lens (n=2.5)and could resolve the striae
>>perfectly well as shown at
>>http://www.mikroskopie-forum.de/read.php?2,25394,25394#msg-25394 .
Your response to that quote:
>I'm not sure exactly where your exp did go wrong. By 'direct
>projection'  you mean you simply took out the projection lens/eyepiece
[quoted text clipped - 3 lines]
>place where the secundary image is placed (where your eyepiece 'picks
>it up'), somewhere a cm or so below the rim of the tube).

You say you are not sure where my experiment went wrong. Well, it
just simply did not go wrong! The image was taken by placing the CCD
of the mentioned camera to exactly where the intermediate image is
located and taking the picture. The picture itself was only 100x  (the
magnification of the objective) x 70 micrometers( the length of the diatom),
or 0.7 cm tall. This is smaller than the vertical dimension of the CCD leading
to a loss of contrast and inability to see the striae by the mechanism I have
described above.

It may well be that you did not read my notes well. In these notes it
is stated that I am using an Olympus BX51, a microscope with infinity
optics and a fully corrected intermediate image. This intermediate
image is accessible above the dovetail connection on the trinocular
leading to the camera port.

You quote a prior post of mine:
>> Finally I wish to point out that my image of A. pellucida (see
>> the link in my original post) was obtained using a magnification
>> of 250x (100x from the objective, 2.5x from the projection lens),
>> while Dr. von Heurck used magnifications of 1800x to 3000x.
Your response to it:
>That would make your specimen a very big one ;-)

Please see the arithmetic I have gone through above, showing that
my diatom was of regular dimension and not a "very big one".

From your recent post:
>ps, keep it up, you are learning a lot.

Rene, please refer to the first paragraph in my response.

G.R.
rene - 29 Jan 2007 16:47 GMT
GR, I still find you lacking on all points except for the Baertierchen
info.

>You seem confused by the issue of contrast. I will say it
> one more, final time: contrast increases when the target's
> image spatial frquencies move to the left on the MTF curve.
> The only way for this to happen at a fixed pixel or grain size is
> to decrease the spatial frequency, that is to increase the size of
> the image.

So what you are suggesting is that if we enlarge the diatom, then we
are able to see more. Duh.

René.
GTO - 31 Jan 2007 02:19 GMT
Hello, Rene and GR:

I am getting headache from reading this part of the thread. I am convinced
that both of you know how to succeed in photomicrography, but this part of
the thread entered the state of verbal jousting and looping.

I can only hope that both agree on the following basic summary:

1) Smallest resolvable distance 'd' (diameter of Airy disk) is given by d =
0.61*lambda/NA(obj) assuming we are using a NA(cond) of 1.4 and hence get no
limitation due to insufficient illumination.

2) The size of the smallest resolvable distance 'D' in the INTERMEDIATE
image (also referred to as real image) is given by D = d * magnification of
objective

3) To resolve a feature of the size 'D', one needs a pixel size of D/2
(Nyquist theorem [1]) to avoid undersampling. But remember, this is the
ACTIVE pixel size and not the PHYSICAL pixel size in case of a Bayer filter
(see point 4). Let's call this the required pixel size 'r'.

4) The required pixel size for a Bayer filter depends now on the wavelength.
If blue light is used the active pixel size is 2 times the physical pixel
size. For green, it is square-root 2 times the physical pixel size. Let's
call a = D/2 to be the active pixel size.

5) To be able to resolve the entire image, the active pixel size must be
smaller or equal to the required pixel size: a <= r

Now let's look at an example:

For a Nikon D70 (I don't know the Pentax that GR is using), the active pixel
size 'a' is a HUGE 12um when we use primarily blue light (no green). This is
much too big to resolve the fine structure with the CCD of a D70 located in
the intermediate image plane (direct projection). For an objective with NA
1.40 with lambda = 425nm, d = 0.185 and D = 18.5 um. This gives r = 9.25 um,
which is smaller than the 12um provided by the D70 for blue light. When
using a 2x projection ocular (or relay lens) the image information is spread
out onto a larger area and hence we can multiply 18.5 um by two = 37 um.
Now, r = 18.5um, which is easily captured by the image sensor's pixels with
a PHYSICAL size of 6um and using blue light.

Now that would explain one aspect of the fact that GR cannot resolve AP when
using direct projection but by using a relay lens. Another important aspect
is, of course, the fact that using the appropriate relay lens, one gets a
complete correction for aberrations when using older gears.

Cheers,

Gregor

[1] http://en.wikipedia.org/wiki/Nyquist-Shannon_sampling_theorem

GR, I still find you lacking on all points except for the Baertierchen
info.

>You seem confused by the issue of contrast. I will say it
> one more, final time: contrast increases when the target's
> image spatial frquencies move to the left on the MTF curve.
> The only way for this to happen at a fixed pixel or grain size is
> to decrease the spatial frequency, that is to increase the size of
> the image.

So what you are suggesting is that if we enlarge the diatom, then we
are able to see more. Duh.

René.
GTO - 31 Jan 2007 02:47 GMT
Please change point 4 into:

4) The pixel size for a Bayer filter depends now on the wavelength.
If blue light is used, the active pixel size would be 2 times the physical
pixel
size. For green, it is square-root of 2 times the physical pixel size. Let's
call the active pixel size 'a'.
NoSpam - 22 Jan 2007 21:29 GMT
Hallo,

I have now had time to follow up on the very valuable
suggestion to look at Edmund J. Spitta's book on "Photo-
micrography" of 1899 and search it for A. pellucida de-
scription and image.

The book can be found at
http://books.google.com/books?vid=OCLC03895698&id=bNYKAAAAIAAJ&pg=RA5-PA1&lp
g=RA5-PA1&dq=spitta+edmund
and downloaded from that page in pdf format
without charge.

As the search function for this type of Adobe pdf page
does not function I am giving the pages to enter in order
to go directly to A. pellucida description and figure.

The description of how the image in this work has been
taken starts on page 163. How others have done this work
is also described. The table of contents for Plate 4 in which
one image of A. pellucida is presented is on page 199, while
the Plate itself with Fig. 4 showing A. pellucida is found
on page 201.

This picture shows striae. No picture showing dots is pre-
sented. The author discusses the possibility that dots if
obtained may be an artifact.

The details for the picture showing A. pellucida with striae
are as follows:
Distance between striae :100,000/inch (that is 0.254 microns).
Zeiss Apo 1.40; Oculoart 18; Oblique light; Crescent shaped
Diaphragm.

I hope you will find this information of interest.

GR.

> Hi
>
[quoted text clipped - 19 lines]
>
> David
justbeats - 25 Jan 2007 18:50 GMT
Great thread!

Resolving AP ain't that tough (with enough NA). Doing so with
sufficient contrast to see it, is!

DIC has best resolution in one direction (orthogonal to the plane of
polarisation I think). I don't know if that's slightly greater than the
equivalent brightfield or not (as someone suggested). I have seen the
pores of AP appear and disappear at 90 degree intervals as the
frustrule is rotated.

Reckon Mr Kemp might get a few orders for AP 'cos of this thread ;-)
Might grab some mounted with no coverslip to try a few epi techniques.
Hadn't thought of that before.

NoSpam mentioned "plugged pores" which triggered a thought. Biological
sections are stained to increase contrast so could the pores (only) of
AP be plugged with an opaque material? They'd stand out in bright field
then! Or embed (cast?) them on a slide with that material and treat
like a rock section; done before?

   Cheers
   Beats

> Microscope Group,
>
[quoted text clipped - 105 lines]
> Regards
> G.R.
NoSpam - 25 Jan 2007 21:38 GMT
Steve,

You touch on important points.

First is the effect of the direction of the diatom's axis with respect to
the
direction of  polarization upon achievable resolution in DIC observation.
In my case the diatom was as right angles to the direction of the polarizer
and parallel to the direction of the analyzer and at 45 degrees to the so-
called shear axis. The image and pertinent text is at:
http://www.mikroskopie-forum.de/read.php?2,25394,25394#msg-25394 .

I did try at one time to rotate the specimen, but this lead to loss of focus
and loss of good immersion oil contact.

Regarding the pores I have the following to say. At
http://www.mikroskopie-forum.de/read.php?2,25440,25440#msg-25440
I reproduce TEM-images from the article "Fine structure of the Diatom
Amphopleura Pellucida" by E.F. Stoermer and H.S. Pankratz from the
Amer. Jour. Bot. 51(9): 986 - 990, 1964.

This article describes the punctae (pores or dots or pearls) as a series of
rounded openings of  60 millimicrons in diameter which in turn are
covered by porous plates with even finer pores of some 6 to 8  milli-
microns in diameter. I hypothesize that if these porous plates are not
removed by the cleaning process, they may obstruct the imaging of the
punctae (pores,dots, pearls) by light microscopy. The tiny pores in the
porous plates may not even be visible on SEM if the sputtering process
obscures them. Then they may also obscure the punctae on SEM.
Obscured pores are visible in some of Mr. K. Kemp's SEM samples.
It may be important to verify the existence and nature of these porous
plates.

I recommend a study of the mentioned article or at least of the pictures
in the above post of mine. One picture shows one horizontal row of porous
plates, while the other image shows the rectangular arrangement of the
punctae covered by porous plates.

Note that the 60 millimicrons quoted for the diameter of the punctae by
Stoermer and Pankratz does approximately, but not precisely agree with
Dr. Gaugler's determinations in the image
http://photoweb.net/micro/Apellucida-I-50kx-1.jpg .
Other images from Dr. Gaugler's FESEM are shown at
http://photoweb.net/micro/ .

I hope the forgoing is of interest to you and the group.

G.R.

> Great thread!
>
[quoted text clipped - 129 lines]
> > Regards
> > G.R.
justbeats - 25 Jan 2007 23:00 GMT
The TEM images are very interesting. Given the irregularities of the
pores, the absolute, perfect regularity of the finer pores is
surprising - almost suspicious...

> Steve,
>
[quoted text clipped - 175 lines]
> > > Regards
> > > G.R.- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -
justbeats - 25 Jan 2007 23:07 GMT
Thought so. I think the nature of those smaller pores is that they are
a halftone pattern. They're the same size in both images!!!

> The TEM images are very interesting. Given the irregularities of the
> pores, the absolute, perfect regularity of the finer pores is
[quoted text clipped - 179 lines]
> > > > Regards
> > > > G.R.- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -
NoSpam - 25 Jan 2007 23:12 GMT
Steve,

Keep in mind that the image is a digital camera copy of an
image transmitted from a digital storage device for rare
journals to a printer. One would have to see the original article
to be sure whether the images are an artifact. They were not
thought to be an artifact by the authors of the article.

G.R.

> Thought so. I think the nature of those smaller pores is that they are
> a halftone pattern. They're the same size in both images!!!
[quoted text clipped - 182 lines]
> > > > > Regards
> > > > > G.R.- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted text -- Show quoted text -- Hide quoted
text -- Show quoted text -- Hide quoted text -- Show quoted text -
rene - 29 Jan 2007 16:41 GMT
Hi Steve, no it's real, see for example my micscape article on
Pseudonitzschia. Of course, this membrane like material covering the
pores is so thin, it wouldn't be visible in any LM image. It's on
molecular basis, the poresize is set by the assembling machine,
therefore very regular.

Rene.

> Thought so. I think the nature of those smaller pores is that they are
> a halftone pattern. They're the same size in both images!!!
Gary G - 28 Jan 2007 19:11 GMT
I believe that I have found a definitive reason why A. pellucida is so
difficult to image using LM.  When I looked at the inside (concave)
side of the diatom, the pores are about 100nm x 145nm at a pitch of
285nm.  With deep UV, one should be able to resolve one pore from
another.  However, resolving the pore itself would not be possible.

Now, with new specimens from Klaus Kemp of the outside (convex) of
diatoms, I find new info.  These pores are about 30nm x 100nm and are
90 degrees rotated from the inside pores.  Basically, the pores look
like:

http://www.photoweb.net/micro/Apellucida-pore.pdf

Thus, if one was trying to see the large pores, the light is
restricted by the underlying small pores.  If one was trying to
resolve the small pores, that already is a no go.

http://www.photoweb.net/micro/Apellucida-100kx-1.jpg

already showed this.  Look at the lower right area and one can see the
90 degree difference of inside and outside pores.

gg

Kiss French.  Drink California.
 
Sign In
Join
My Latest Posts
My Monitored Threads
My Blog
My Photo Gallery
My Profile
My Homepage

Start New Thread
Enable EMail Alerts
Rate this Thread



©2009 Advenet LLC   Privacy Policy - Terms of Use
This website includes both content owned or controlled by Advenet as well as content owned or controlled by third parties.