Yes, I know, it's for metallurgy! But how do the optical properties of
these objectives differ from normal brightfield ones? How might they
be expected to perform for biological applications, for example?
metalengr@hotmail.com - 21 Oct 2007 19:42 GMT
On Oct 21, 11:31 am, UKOncol...@aol.com wrote:
> Yes, I know, it's for metallurgy! But how do the optical properties of
> these objectives differ from normal brightfield ones? How might they
> be expected to perform for biological applications, for example?
Metallurgical objectives are meant to be used with reflected light and
without a cover slip (or cover glass).
The optical tube length is longer (typically 170mm or even 210 mm)
versus 160mm in order to allow room for the vertical illuminator.
Therefore they would be expected to deliver lower performance than
biological objectives.
Pittsburgh Pete
DISCLAIMER
We do not believe what we write, and neither should you. Information
furnished to you is for topical (external) use only. This information
may not be worth any more than either a groundhog turd, or what you
paid for it (nothing). The author may not even have been either sane
or sober when he wrote it down. Do not worry, be happy.
Kevin Cunningham - 21 Oct 2007 22:59 GMT
On Oct 21, 1:31 pm, UKOncol...@aol.com wrote:
> Yes, I know, it's for metallurgy! But how do the optical properties of
> these objectives differ from normal brightfield ones? How might they
> be expected to perform for biological applications, for example?
Pete's right, mostly the difference is that met objectives don't use
cover glasses. However no one uses tube length any more. Now every
thing is infinity corrected. Another big difference is now all met
microscopes have standardized magnifications at 5, 10, 20, 50 and 100
times.
The quick answer is usually met objectives won't work for biologically
prepared specimens.
Thanks,
Kevin Cunningham
SMS
mc - 22 Oct 2007 00:19 GMT
> The quick answer is usually met objectives won't work for biologically
> prepared specimens.
Well, they work, and you might not notice that you're not getting top
quality, but a conventional objective that works better is less expensive.
justbeats - 25 Oct 2007 22:40 GMT
On 21 Oct, 18:31, UKOncol...@aol.com wrote:
> Yes, I know, it's for metallurgy! But how do the optical properties of
> these objectives differ from normal brightfield ones? How might they
> be expected to perform for biological applications, for example?
Assuming epi-objectives are "metallurgical objectives", I find the
darkfield ones work well in some niche applications. I got nice
closeups of stomata by stacking shots in HeliconFocus. Brightfield
doesn't work too well visually - far too much glare - but you can get
reasonable results photographically with heavy processing. All of this
on fresh leaves or wood etc - no coverslip...
heliogabalus - 26 Oct 2007 00:39 GMT
> On 21 Oct, 18:31, UKOncol...@aol.com wrote:
> I got nice
> closeups of stomata by stacking shots in HeliconFocus.
when i try to assemble the images with helicon the single frames are
dislocated, so the final image presents borders of the images below
can i know the reason of it?
justbeats - 26 Oct 2007 14:00 GMT
> > On 21 Oct, 18:31, UKOncol...@aol.com wrote:
> > I got nice
[quoted text clipped - 4 lines]
>
> can i know the reason of it?
Couple of possibilities...
1. You are not using enough images and your individual focus planes
are too far apart. Take more pictures more closely spaced (in z), that
definitely solves the "stairstep edges" problem - which is most likely
what you're seeing...
2. In the options, ensure the ability to shift the image left and
right during alignment is enabled. However, I always turn off the
option to allow rotation for alignment as that sometimes does the
wrong thing and leaves edge artefacts
Hope that helps...
heliogabalus - 27 Oct 2007 20:45 GMT
>> > On 21 Oct, 18:31, UKOncol...@aol.com wrote:
>> > I got nice
[quoted text clipped - 18 lines]
>
> Hope that helps...
yes , that helped : )