In the good ole days, Kohler illumination was done with a lamp who's
filament was about as wide as your finger. However your just not
going to find any of those lamps.
A ground glass filter should be used, I'll bet there is a place for
one. Other wise I'd put it close to the lamp, if possible directly in
front of the lamp. Then, technically, your using critical
illumination, not Kohler but who cares?
Thanks,
Kevin Cunningham
SMS
NoSpam - 28 Nov 2007 18:45 GMT
Dear Kevin,
Thank you for your helpful reply. There is only
one easy way to insert a diffuser and that is in a pre-
mounted swing-out filter holder between the field lens
and the condenser. Any other approach would require
mechanical modifications to the scope.
A diffuser in the filter holder would work fine, provided
it is fine grained and does not rob too much light. The
filter holder accepts round glass 33mm in diameter.
Do you have any idea where to get a suitable diffuser
with this dimension?
Thank you again
GR.
> > Hello,
> >
[quoted text clipped - 42 lines]
> Kevin Cunningham
> SMS
NoSpam - 28 Nov 2007 19:06 GMT
Dear Kevin,
I forgot to raise a point: Is opal glass or ground glass the
choice for a diffuser?
TIA
GR.
Kevin Cunningham - 28 Nov 2007 20:35 GMT
> Dear Kevin,
>
[quoted text clipped - 3 lines]
> TIA
> GR.
It depends on how much diffusion and for what purpose. In this case
ground glass would be better. This is one of those "Don't get me
started" kinda things. If you want the next couple of hours filled up
with discussion of who, what and how many I can do it, if you want to
make your Nikon work, ground glass.
Thanks,
Kevin Cunningham
SMS
Richard J Kinch - 29 Nov 2007 08:18 GMT
> I forgot to raise a point: Is opal glass or ground glass the
> choice for a diffuser?
http://www.luminitco.com/
NoSpam - 29 Nov 2007 16:24 GMT
> > I forgot to raise a point: Is opal glass or ground glass the
> > choice for a diffuser?
>
> http://www.luminitco.com/
Dear Dr. Kinch,
Most interesting! I have looked at Luminit's pdf on Light
Shaping Diffusers (LSDs) and found a description of such
a diffuser which exactly fits my needs.
A pertinent image and text can be found under "Homogenized
Light". The advantage of their diffusers seems to be that they
cause very little loss of light.
There is of course a question not answered by this brochure:
Is the probably very high price of Luminit's LSDs justified by
the gain they make in light throughput compared to cheap
ground glass?
Kevin is right: modern scopes all use light diffusers to hide the
filament structure of the illuminating lamp. I wonder whether
Luminit like diffusers are used in these scopes.
Another interesting fact seems to be that centering screws for
the condenser no longer are needed and provided. I wonder
whether it has been found that the mounting can be manufactured
accurately enough to avoid the centering mechanism altogether.
Fact is that modern and expensive scopes no longer offer what
is known as Koehler illumination.
Thnanks for your interest.
GR.
Richard J Kinch - 30 Nov 2007 20:56 GMT
> There is of course a question not answered by this brochure:
> Is the probably very high price of Luminit's LSDs justified by
> the gain they make in light throughput compared to cheap
> ground glass?
Well, if you just need one, they'll be glad to send you a free sample.
They're used in a lot of consumer gadgets, so I don't think they're that
expensive.
NoSpam - 30 Nov 2007 21:58 GMT
> > There is of course a question not answered by this brochure:
> > Is the probably very high price of Luminit's LSDs justified by
[quoted text clipped - 5 lines]
> They're used in a lot of consumer gadgets, so I don't think they're that
> expensive.
I hope you guessed right. I already asked them for a free sample. We
shall see whether they are generous enough to part with a sample.
GR..
Kevin Cunningham - 28 Nov 2007 20:40 GMT
> Dear Kevin,
>
[quoted text clipped - 59 lines]
> > Kevin Cunningham
> > SMS
I'd contact Nikon first, they Olympus of course the microscope div.
In the back of the Olympus price list in the old days there used to be
tons of filters. They both use lots of ground glass filters and one
should have the size you need. If not try an on line photo shop
although, this is my personal view, you don't need less resolution in
any camera.
Thanks,
Kevin Cunningham
SMS
Andy Resnick - 28 Nov 2007 20:34 GMT
<snip>
> A ground glass filter should be used, I'll bet there is a place for
> one. Other wise I'd put it close to the lamp, if possible directly in
> front of the lamp. Then, technically, your using critical
> illumination, not Kohler but who cares?
I don't think that's correct- critical illumination is the Fourier
transform of Kohler illumination: the filament is focused onto the
sample plane rather than the entrance pupil of the condenser. And
actually- that's the better solution.
Using a ground glass plate is a kludge to destroy the image of the
filament in the back aperture of the objective lens; it may work, but it
isn't critical illumination.

Signature
Andrew Resnick, Ph.D.
Department of Physiology and Biophysics
Case Western Reserve University
NoSpam - 28 Nov 2007 22:36 GMT
> <snip>
>
[quoted text clipped - 11 lines]
> filament in the back aperture of the objective lens; it may work, but it
> isn't critical illumination.
Hello Andrew,
Thanks for your clarification. It would be interesting to try critical
illumination. Unfortunately I can't think of a way to place the light
source or its image into the plane of the field diaphragm on my
Nikon scope. Should there be a way I would be happy to try it.
Please let me know.
G.R.
Andy Resnick - 29 Nov 2007 21:01 GMT
>><snip>
>>
[quoted text clipped - 19 lines]
> Nikon scope. Should there be a way I would be happy to try it.
> Please let me know.
If your source has a focussing knob, you could try using that. In all
honesty, critical illumination is rarely used, because when the image of
the source is superimposed on the sample, viewing the sample means
looking at a magnified image of the bulb filament.

Signature
Andrew Resnick, Ph.D.
Department of Physiology and Biophysics
Case Western Reserve University
Kevin Cunningham - 29 Nov 2007 13:14 GMT
> <snip>
>
[quoted text clipped - 16 lines]
> Department of Physiology and Biophysics
> Case Western Reserve University
Dr. Resnick, Good points! I've been having this discussion over a
few years with some buddies in Germany and Austria. I should have
known better. What NoSpam and I were discussing was whats commonly
called either critical or Kohler illumination.
In fact no one uses either exactly. The usual start point is a ground
glass assembly, take a look in the base of a modern microscope there
are between 1 and 3 ground glasses. This is because designers know
that if you use a specially designed lamp with a thick filament they
are a pain to get, expensive and in a few years very few will be
used. Its simpler to use ground glass and a lot cheaper.
Thanks,
Kevin Cunningham
SMS
Andy Resnick - 29 Nov 2007 20:58 GMT
>><snip>
>>
[quoted text clipped - 28 lines]
> are a pain to get, expensive and in a few years very few will be
> used. Its simpler to use ground glass and a lot cheaper.
For home use, I sort of agree with you- a diffuser allows a hobbyist to
quickly and easily quasi-align the microscope for even illumination.
I think it's important to remember that Kohler/critical illumination is
more than just simply focussing and centering the condenser field stop
onto the sample- it means that all the aperture and field stops
(condenser and objective) are at Fourier transform planes, and that the
source is also at one of those planes. Which plane the source is at
(field planes or pupil/aperture planes) determine if the microscope is
"Kohler" or "critical".
In the lab, I remove the diffuser (Leica uses some wierd diffraction
element as opposed to a ground glass plate). Our users don't use
conoscopic viewing, so images of the filament are not really a problem.
An alternative solution I use is to pipe the halogen light in with a
liquid-core fiber.