Gregor,
Thanks for your answers.
My situation is that I already had a Zeiss standard and bought, over the
years, some planapo's.
Recently I bought a Nikon photomicroscope and I like it very much. For
microphotography (or photomicrography, whatever you want) it is great. So I
tried my Zeiss objectives in it and got bad results. Then I tried my Zeiss
objectives plus Zeiss Kpl eyepieces and got good results. Obviously
objectives and eyepieces of a specific brand and type are matched.
But are the condensors matched, too? Is the illumination from a Zeiss
standard the same as the illumination of my Nikon?
I also read about high-NA aplanatic and/or achromatic condensors. Will these
give better results (apart from the fact that I never saw one for sale ...)?
Are they difficult to use?
Theo
1) From http://www.microscopy-uk.org.uk/mag/artmar04/godarkfield.html,
the limit of resolving power for a compound microscope is given as
d[µm] = 1.22 lambda[µm] / (NA_Condenser + NA_Objective)
Hence, this will depend on what high-NA planapos you are talking about.
2) Don't! Espcially, don't waste your time with this if you plan on
using oclulars from Nikons CF and CFI60 system. Make a google search
for "Chromatic Difference in Magnification" (CDM).
3) Yes! See point 1 above.
4) No. This only applies if there is no specimen in the optical path
and you just want to illuminate the entire backfocal plane of the
objective. Otherwise, diffraction in the specimen plane "boosts" the
resolution capability of your setup (see equation in point 1 above).
5) Clear since previous point is not true.
Gregor
GTO - 30 Jun 2005 00:41 GMT
Theo:
This will depend on your application whether you should use an
aplanatic high-NA condenser (with NA 1.4). For my work, I usually use
an Achromatic condenser (dry) with swing-out top at NA 0.9. It works
great for most of my brightfield work. For oil, I still use a simple
Abbe top with NA 1.25.
Mixing condensers can be tricky. For instance, the better condensers
for the Nikon E400 will not allow you proper Koehler illumination on a
Nikon E200. The phase condenser from a Nikon system does not work on a
Leitz Ortholux. You have to be very careful when mixing components.
Especially when you want decent Koehler illumination for your work.
Gregor
Gordon Couger - 30 Jun 2005 07:10 GMT
> Gregor,
>
[quoted text clipped - 13 lines]
> give better results (apart from the fact that I never saw one for sale ...)?
> Are they difficult to use?
Hi Theo,
As you have discovered older Zeiss objectives have part of the
correction built in the Eyepiece.
The same is true below the stage all the parts need to work
together to light the subject properly. If you use a mirror and
old AO Spencer focusing lamp with a diaphragm you can make any
condenser work. But if you use the lighting built in the
microscope you don't have enough adjustment to make every
condenser work on ever microscope.
The above and blow stage part don't need to be the same brand
except phase rings must match DIC prisms must work together and
any other thing that has parts on both sides of the subject
generally need to match.
Gordon