>I'm Using an EOS 10D mounted on the front photoport of a Zeiss ICM
>405. As I have posted before, this setup has FOV issues caused by the
>small sensor (15.1mm x 22.7mm) compared to the 35mm film the ICM 405 is
>designed for; I live with this and do more mosaicing on specimens that
>need it - but it would be nice to have a fix.
I'm not familiar with the ICM (?); I use an Ultraphot 3 for most of my
photomicroscopy.
>One solution I considered was to buy a DSLR with a 35mm sensor (1D or
>1DS)
Note the 1D does not have a full-frame 35mm sized sensor - it has a 1.3x
factor rather than the 1.6x of the 10D/20D/300D/350D series though.
> but this is VERY expensive option and (worse) because the pixels
>on this sensor are EVEN smaller than those on the 10D, it exacerbates
>another issue I've been hitting. Namely; small pixels result in gross
>over-sampling - which is the intended topic of my post...
One thought - if you could find a second hand 1Ds (Mk 1 version) with an
11 MP full-frame sensor, you could save a lot of money and reduce the
oversampling.
>I'm interested to hear others comments and/or experiences on this.
>
[quoted text clipped - 32 lines]
>used. This state of affairs wastes pixels, download time (from the
>camera) and processing power (on resulting images).
You would know your own needs best, of course - but is the oversampling
^that^ much of a problem? You can surely just downsample the files
before further processing if necessary.
>Possible solution #1
>--------------------
>Bigger camera pixels. New camera needed, therefore expensive and maybe
>not one available with the ~20um pixels needed anyway. Not attractive.
One of the MF digital backs (does your ICM take such a thing). A
second-hand one with lower pixel count should suit. Still very expensive
though.
>Possible solution #2
>--------------------
>Find a gizmo to reduce image scale (e.g. a 0.5x focal reducer). Tried
>to find one to solve the FOV problem and failed - even after getting
>hold of a knowledgeable tech at Zeiss Germany. Maybe I missed a product
>somewhere?
Again, as I am not familiar wit the ICM, this may not suit; however,
there is a wide-field 5-position changer which has a 0.63x magnification
factor. My specimen does not have a part number visible, but from the
catalogue it appears to be 47 31 55.
0.63x is of course 1/1.587, so it should almost exactly cancel out the
1.6x factor of the sensor. It would however not allow rapid switches
between the 0.63x and normal (1x) FoV.
>Possible solution #3
>--------------------
[quoted text clipped - 9 lines]
>Get higher NA objectives. I would, but I don't think they're
>available and the cost would be prohibitive anyway.
Assuming your stand is a 160mm tube length one, the 63x NA 1.4 is the
only one which springs to mind. I paid about £500 for mine. Wouldn't
make that much difference, though it ^is^ a superb lens.
>Possible solution #5
>--------------------
[quoted text clipped - 9 lines]
>--------------------
>I have no possible solution #6...
(6) Try using a fixed-lens digicam? In addition to my 10D, I also use a
Coolpix 995 fitted directly to a Zeiss KPl 10x20 eyepiece (screwcut to
fit) which slides directly into an eyepiece tube. I'm sure something
like this could be adapted to fit, though I don't know the configuration
of your photoport. The 3 MP would probably be sufficient, and using the
zoom control would give you variable FoV. You also get real-time
focussing output from the camera. The later version (4500) has 4 MP;
later Coolpix models have much wider lens fronts and are harder (but not
impossible) to fit.
(7) Get a dedicated photomicroscopy camera - see Gary Gaugler's recent
post.
(8) Try a webcam and lots of stacking.
Hope these thoughts help - if your ICM is a modern infinity system most
of them may be irrelevant. If we can clarify further, some details of
the scope would help.
David

Signature
David Littlewood
I would not say this. It is very likely that you are just using the wrong
relay lens for your setup. Take one that is suitable for your scope. What
scope are you using anyway? - I am using a D70 with 6MP on a Nikon Eclipse
scope. The D70 has a 1.5x cropping factor compared to the "regular" 35mm SLR
format. Can you see over-sampling in the following JPEG output:
http://www.pbase.com/overney/image/39648499/original.jpg
Before I would buy a 35mm sized DSLR for a scope, I would spend the money on
a dedicated camera with CCD cooling. Nothing beats this in quality, ease of
use and low noise. A DSLR on a scope makes only sense if you use it also for
other types of photography. Otherwise, it's a waste of money and much too
tedious to make it work. Compared to a Coolpix 5000, the quality is
marginally better. Compared to a Coolpix 990, it is better. Compared to a
SBIG cooled CCD (even with just with 2 MPixels), the DSLR is worse for low
light application. However, the DSLR is better when used with a 4x APO lens
with an NA of 0.20 than any dedicated camera with less than 4MPixels but
with a Bayer filter.
Gregor
> I'm Using an EOS 10D mounted on the front photoport of a Zeiss ICM
> 405. As I have posted before, this setup has FOV issues caused by the
[quoted text clipped - 84 lines]
> --------------------
> I have no possible solution #6...
justbeats - 27 Jul 2005 08:49 GMT
> Can you see over-sampling in the following JPEG output:
Well, no - but unless it's a crop out of a larger image, this pic isn't
at the full resolution of your D70 - it's only around 1500 by 1100
pixels. Isn't the D70 capable of something around 3k by 2k?
Clarification - I am not using a relay lens (other than any that might
be built into the internals of the ICM405). I'm projecting directly
onto the sensor using the front photoport which is actually designed to
project directly onto 35mm film frames within a dedicated camera back
instead. I have used adaptors to mount the EOS 10D there instead of the
(M35?) Zeiss film camera that came with my system. I use the Zeiss
exposure control box to hold the internal shutter open in bulb mode - I
then use the EOS shutter (with mirror lock) to control exposure times
directly.
What I need (but can't find) is a "relay lens" offering magnification
of 0.6x, but these only seem available for video systems and their tiny
little CCD sensors. I can't find one for my setup...
I have a coolpix 4500 too - I think the EOS 10D produces better
pictures, and offers better control etc.
I also have a megapixel cooled CCD I use for astrophotography. I tried
this for imaging on the microscope first (before purchasing the 10D).
It works, but has even more oversampling as the pixels are 6um compared
to the 7.4um of the 10D. FOV is pitiful. You're right though - this is
unsurpassed for v. low light work...
Anyway - back to the point, while I agree your image doesn't look
oversampled, I also can't argue with the numbers I calculated and
results I am getting (which agree with the numbers). Do you have more
info on your image? What objective, magnification, NA etc? What is the
image scale (i.e. how many pixels/micron in the final image).
Cheers
Beats
GTO - 27 Jul 2005 22:19 GMT
The picture is a crop to be able to compare it with an image the
Coolpix 990 produced. The D70 captures the entire real image with some
vignetting. By removing vignetting, I loose roughly 1MPixel of data and
hence end up with a 5 MPixel shot. The following image shows the entire
FOV but it is then resized to match a practical format for pBase (see
Pine Single Needle Leave c.s. at
http://www.pbase.com/overney/image/44874329.jpg.)
I found that dealing with vignetting is much less painful than trying
to find a 0.6x relay lens that covers adequately the entire CCD of a
DSLR, which has a 1.5x cropping factor. I haven't found one yet either.
(BTW, my scope does no require a relay lens to fully correct for CDM.)
I also do not yet believe Nikon's claim that the market will never
accept 35mm image sensors in DSLRs ("because they would always remain
too expensive and required better corrected lenses, which also add to
the cost of ownership"). So far, Nikon decided to stick with the so
called DX format, while Canon seems to dance on both parties
simultaneously. - Why invest now a huge amount of money into the
perfect setup when we might be able to buy a 35mm image sensor, which
allows us to bin neighboring pixels to avoid "oversampling" and
increase S/N performance?
The details relevant for the image I mentioned in my previous message
are:
- D70 in bulb mode using IR remote and external shutter (very similar
to your setup). I built the external shutter myself. It's primitive but
it works well down to ¼ second and for flash sync.
- Nikon Plan FLUOR 40x with NA 0.75 (corrected for up to 25mm FOV)
- Nikon E400 scope
- Koehler illumination using brightfield with an Achromatic condenser
NA 0.90
- Trinocular setup (of course)
- I use Nikon's RAW format (NEF)
- Vignetting is cropped away using batch processing using Nikon Capture
4.3 software
- Image is resized and cropped to match the output of the Coolpix 990
(see http://www.pbase.com/overney/technology_photomicrography for a
comparison between Coolpix and D70)
- JPEG generation is done using "normal" sharpening with Nikon Capture
4.3 software
Regarding your question whether the D70 has a true 3k x 2k resolution,
I must say no. It has a Bayer filter mounted on top of the CCD and
hence it requires some fudging in firmware to recreate a 3k x 2k image
with the correct intensities for each pixel. Such an approach is
noticeably worse than a 4MPixel camera using three CCD image sensors.
But I agree with you that even when using a Bayer filter, the D70's
resolution is enough to sample the entire FOV when I use a 40x
objective with an NA of 0.75.
Cheers,
Gregor