When we take away the emission filter and look via the eyepieces we see
2 different spots of our laser light.
Description of our optical path:
We focus our 405nm laser beam through a dichroic mirror which reflects
the beam to the back focal plane of an Olympus PLAPON60XO/TIRFM oil
objective (we want illuminate exit samples with a collimated beam).
These are reflections and not fluorescence spots - using a 406/15
emission filter did not change this phenomenon. We had also similar
observations with some other objectives. It occurs both if we put
immersion oil on the objective and also without it. And of course no
sample is on the objective in that test.
Did anybody come across the same problem using similar wavelength and
using that Objective? Did someone here have other similar observations?
Should we see any spot at all if the system is "fine"?
Lots of thanks,
Adi
Kevin Cunningham - 30 Mar 2006 13:56 GMT
> When we take away the emission filter and look via the eyepieces we see
> 2 different spots of our laser light.
[quoted text clipped - 15 lines]
> Lots of thanks,
> Adi
This is just a thought only. You may be getting a reflection from either
the top lens of the objective or from something under the stage like a
condenser. I've seen the same thing but from fluorescence microscopes.
Kevin Cunningham
SMS
justbeats - 30 Mar 2006 19:55 GMT
Probably nothing to do with your problem, but interesting to note that
"dichroic" has two distinct meaning in optics. Polarisation involved in
both cases...
1) selection or absorbtion by wavelength (your application)
2) splitting of beam paths by wavelength (your problem - maybe?)
Uncanny...