helium gas flow cryostat tips?

Thread view: 
Enable EMail Alerts  Start New Thread
Thread rating: 
dq - 16 Feb 2008 22:23 GMT
Hi, I tried this in s.p.condensed-matter, but maybe I'm better off here
because this group is actually active.  Here we go...

So if you are doing microscopy though a gas flow-cryostat and it looks
like your subject is below the surface of a pond, it likely means you
have liquid helium flowing over your sample when you should really only
be getting cold gas helium.

I think the best way to avoid this is to choke off the helium flow a
little and turn up the resistive heat slightly, to make sure you boil
off the incoming helium before it gets into your optical path.  I tried
this today and built up enough pressure to blow out the sample insert
out of its position (luckily, nothing was damaged).

How do I avoid this?  My guess is that I need to make sure to get the
helium flow down to a minimum before I go heating up the element.

Does anyone have any other cryo tips?  There's precious little
literature on this subject, so I'm making an appeal to the great usenet.
sidd@situ.com - 17 Feb 2008 08:01 GMT
snip--

>So if you are doing microscopy though a gas flow-cryostat and it looks
>like your subject is below the surface of a pond, it likely means you
>have liquid helium flowing over your sample when you should really only
>be getting cold gas helium.

what is the cryostat ? ie  are you taking cold gas off a dewar and
putting it into the cell? or introducing liquid from a dewar into
an intermediate chamber, evaporating it there before the gas
goes into the optical cell ?

i think you will have a factor of 800 or so in volume expansion if you
turn up the heat to boiling. what it your helium consummption ?
with that and the latent heat of evaporation (20 J/g if i recall)
you should be able to calculate a heat input ...

>I think the best way to avoid this is to choke off the helium flow a

are you introducing liquid ?
have you a needle valve ?
expansion valve ?
is the cell at atmospheric pressure or is it pumped down ?

>little and turn up the resistive heat slightly, to make sure you boil
>off the incoming helium before it gets into your optical path.  I tried

sounds like you need less heat, and a pressure gauge in the cell

>this today and built up enough pressure to blow out the sample insert
>out of its position (luckily, nothing was damaged).

heehee

>How do I avoid this?  My guess is that I need to make sure to get the
>helium flow down to a minimum before I go heating up the element.

watch your thermometers, when there is no liquid in the cell the
thermometer (you have one ?) will respond very quickly to heat
input; when there is liquid it will hardly change at all and
will sit at 4.2K (at atm. pressure)

>Does anyone have any other cryo tips?  There's precious little
>literature on this subject, so I'm making an appeal to the great usenet.

see if you can get the Lounaasama book  or the cornell book
on experimental methods in low temperature fizix
the first reference is primarily for 1K or lower, but has
all maner of data for crypgenics.

sidd
daniel - 18 Feb 2008 21:57 GMT
>> So if you are doing microscopy though a gas flow-cryostat and it looks
>> like your subject is below the surface of a pond, it likely means you
[quoted text clipped - 4 lines]
> an intermediate chamber, evaporating it there before the gas
> goes into the optical cell ?
-cut-
> are you introducing liquid ?
> have you a needle valve ?
> expansion valve ?
> is the cell at atmospheric pressure or is it pumped down ?

Well, from what I can tell (I inherited this lab from a doctoral student
who ran off after giving me only a cursory explanation of how the lab
worked), we're working with liquid helium that should evaporate before
it hits the cell.

Liquid helium enters the cryostat via a coaxial transfer arm and needle
valve.  We pump the return helium out via the same coaxial arm (this
time on the outside), back into a recycling system.  Somewhere along the
line, I think the helium is supposed to evaporate---and before it hits
the sample.  There is no intermediate chamber that I know of.  Is this
possible?  Or do I think I have a cryostat design that was never
invented?  This is an old off-the-shelf Oxford system, so it shouldn't
be anything crazy.

The cell is at atmospheric pressure flooded with warm helium, but the
flowing cold helium should feel the under-pressure from the helium
recycling system pump downstream.  Sometimes the pumping is strong
enough to overpower the evaporation pressure of the boiling helium, and
introduces a vacuum in the dewar.

So to your bullet points:  yes, yes, not that I know of, and it varies.

> i think you will have a factor of 800 or so in volume expansion if you
> turn up the heat to boiling. what it your helium consummption ?
> with that and the latent heat of evaporation (20 J/g if i recall)
> you should be able to calculate a heat input ...

Without heat, I can put away roughly 1 cubic meter of gas helium per
hour.  We measure this directly through the recycling system.  So that's
about 180 grams, or 3600 J, right?

>> little and turn up the resistive heat slightly, to make sure you boil
>> off the incoming helium before it gets into your optical path.  I tried
>
> sounds like you need less heat, and a pressure gauge in the cell

The first part is easy, second part not so much.  I do at least have a
0.05 bar over-pressure safety valve, although it seems like that valve
is cut-off from wherever the pressure builds when the helium boils, hence...

>> this today and built up enough pressure to blow out the sample insert
>> out of its position (luckily, nothing was damaged).
>
> heehee

>> How do I avoid this?  My guess is that I need to make sure to get the
>> helium flow down to a minimum before I go heating up the element.
[quoted text clipped - 3 lines]
> input; when there is liquid it will hardly change at all and
> will sit at 4.2K (at atm. pressure)

There was definitely lag between changing the heater voltage and the
temperature readout.  It wasn't at 4.2K though, more like 6K, which I
think just means the sensor isn't positioned as well as I would like it
to be.

>> Does anyone have any other cryo tips?  There's precious little
>> literature on this subject, so I'm making an appeal to the great usenet.
[quoted text clipped - 3 lines]
> the first reference is primarily for 1K or lower, but has
> all maner of data for crypgenics.

I've only read Ekin "experimental techniques for low-temperature
measurements", which seemed a little too general.  I will see if I can
get a copy of Cornell or Lounaasama.

Thanks for the help.  From what I can tell, I'm not that far off to try
reducing flow and being more careful with the heater...well, back into
the lab I go...
sidd@situ.com - 23 Feb 2008 13:00 GMT
>who ran off after giving me only a cursory explanation

let im get away, did you? find a fone # is my first step

>Liquid helium enters the cryostat via a coaxial transfer arm and needle
>valve.  We pump the return helium out via the same coaxial arm (this
>time on the outside), back into a recycling system.

snip--

>is no intermediate chamber that I know of.  Is this
>possible?  

oh yes

>This is an old off-the-shelf Oxford system, so it shouldn't
>be anything crazy.

well...i might not go quite that far...

>Without heat, I can put away roughly 1 cubic meter of gas helium per
>hour.  We measure this directly through the recycling system.  So that's
>about 180 grams, or 3600 J, right?

mmm i think the 20J/g is for just latent heat to change he4 from
liquid to a (very dense) gas.. then you have the expansion to atmospheric
pressure and temperature which eats many joules so i think you have to put
that in as well

snip--

>There was definitely lag between changing the heater voltage and the
>temperature readout.  It wasn't at 4.2K though, more like 6K, which I
>think just means the sensor isn't positioned as well as I would like it
>to be.

at 6K i tink you are to the right of the (normal) liquid gas
critical point at 5.sumpn in the P-T plane

sidd
 